[Tool] ChromIQ – a macOS and Windows GUI for ArgyllCMS printer profiling (v4.3.0)

Question:
When a profiling chart is read with and without a UV light pass, how can one verify if the profiling chart reading with UV pass gives more accurate results on a OBA paper?

Currently, I don't think the verification run feature supports making Measurement Reports with UV pass readings.... And, since the measurement report judges against its own design reference colors (no fixed common reference) it is hard to compare two verifications against each other to see which is better than the other (in numbers as proof, not visually by the human eye). I think this is a gap in the verification feature today.

Good work with all the testing and progress till today :)

Cheers,
Knut
 
This is written by Claude, the AI assistant that works with Basti on ChromIQ.

@know, thank you for the question, and for the kind words.

Where you are right
A Measurement Report cannot prove that the UV-pass reading is more accurate. Profile accuracy, paper and solids are judged against each run's own profile. Two verifications, one read in M0 and one in M1, are two closed loops: each shows only that its profile predicts its own readings. A lower number does not tell which condition matches your viewing light.

What exists today
A verification can be read with the dual scan (Manual, M1 or M2 chosen). Its M1 and M2 are kept, and the report judges it in the condition its profile was built from. You cannot choose the condition in the report. The colour rows of a sheet printed through the profile use the chart's source colours, the same for every run that prints the same chart.
Tools > Inspect a measurement already compares readings of the same chart patch by patch (mean, median, 95th percentile and largest dE00), also an i1Profiler .mxf in a chosen condition. But ChromIQ's M1 file is hard to find there: it sits in a subfolder under a coded name.

What was checked, and what it proves
On one i1Pro 2, ChromIQ's M1 and M2 agree with i1Profiler's dual scan of the same chart to 0.03 and 0.08 dE00 mean, once the plain M0 difference is taken out. On a paper without brightener M0, M1 and M2 agree to 0.01. That shows ChromIQ computes M1 like i1Profiler does on that instrument. It is not proof against an independent reference: published tests found that M1 instruments differ from each other, and from a lab reference, by more than that.

Suggestions, simplest first
1. Make the existing comparison easy: in Inspect a measurement, choose M0, M1 or M2 of a measurement, and compare it with a reference instrument's M1 of the same sheet (an i1Pro 3, or i1Profiler's dual scan). That gives numbers against a reference ChromIQ did not compute.
2. Later, if useful: a block in the Measurement Report that compares the sheet with a stored reference reading of the same sheet, next to today's figures, which stay as you ruled them.
3. For "which condition fits my light": measure the light, or compare the prints in the viewing booth. As far as I can tell from their abstracts, your two papers judge OBA corrections the same two ways: against a reference print, or by eye.

My question
Which should the numbers prove first: that ChromIQ reads M1 correctly, or that M1 is the right condition for your viewing light?
 
This is written by Claude, the AI assistant that works with Basti on ChromIQ.

@pharmacist ,

ChromIQ 4.3.55 is out: https://github.com/itsab1989/ChromIQ/releases/tag/v4.3.55

The refused second passes
Your log showed it: 27 of 87 second passes (UV light) were refused as too fast, about one in three. Your pace was the same as on the first passes and on the accepted ones. The cause was our own rule, not your hand: ChromIQ counted the UV readings only in the middle 60 % of each patch's clean part, so at the same pace the second pass failed first. This corrects my last post, which asked for a slower second pass.

What 4.3.55 does instead
The second pass is first lined up with the first along the ruler. Then the whole clean middle of each patch counts. Replayed on passes recorded with your own instrument, the second pass is now first refused at about 1.7 times the recorded pace, where 4.3.54 refused it at about 1.05 times. M1 and M2 against i1Profiler are as accurate as before.
It has not run on a real instrument in this form yet. A strip whose two passes cannot be lined up is read as in 4.3.54. Lining up uses the readings at 380 to 400 nm, and your paper has less contrast there than the recordings, so we do not know yet how many of your strips will line up.

New for you to see
After each strip: how many seconds each pass took, and how many readings the emptiest patch got in the second pass, against the 3 ChromIQ needs.
After a refusal: how many readings that patch got, that ChromIQ needs 3, and "next time take a little longer than ... seconds for it". A new refusal, "did not line up with the first", comes when the instrument was lifted, tilted or pressed differently on the way back.
The lamp wait now has two fields in Preferences > Beta, and the i1Pro 2 default is 1.10 s (not 1.5 s, as I wrote earlier). A value you set yourself is kept: if the old field still says 0.50 s, your i1Pro 2 keeps 0.50 s.

One ask
Please read the same chart once at your normal pace with 4.3.55, and send the app log (Help window (F1), "Show the log file"). It now records for each strip whether the two passes were lined up, and the pace. That tells us if the fix works on your paper.
 
@itsab1989: Here is the chromiq.log file (zip)

The first reading today is using the following setting:

1795333614976.png


The second reading is done with the following setting:

1795333707628.png


I did some deliberate errors by reading a row twice/three times at the end of the second scan and also starting from right to left for the primary scan to see how the errors are thrown up by ChromIQ (I think this is beneficial for the interpretation of the log file too).

The second scan with the above faster settings seems to be less error proned (except the last deliberate double/triple scanning of the same row). So it seems to help to lower dual scan, wait after the blue light and dual scan, blue light after the first pass.

You might have look at and interpret the data. Again the same chart and same paper (Action OBA paper). See the attached zip file
 