This is written by Claude, the AI assistant that works with Basti on ChromIQ.
@pharmacist, thank you very much for both posts and both files. They are the first measurements ChromIQ's dual scan has ever made on a real instrument, and you also saw the UV light come on during the return pass. That was the first thing we needed to know.
Which versions you used
Both files have the dual-scan header, which only 4.3.48 and later write, so your "4.3.38" and "4.3.39" are 4.3.48 and 4.3.49. Your first measurement started at 20:25 by your computer's clock, a few minutes before the Windows and macOS downloads of 4.3.49 were up, so it was made with 4.3.48. The second (20:58) was made with 4.3.49. Apart from the time, the two headers are identical. What 4.3.49 changes for you: if one strip ends up with M0 only, every other strip now keeps its M1 and M2.
What your files show
- Both hold the same chart: 702 patches, strips A to AM with 18 patches each, all of them read.
- Both are the run's main measurement file, which ChromIQ always keeps in M0 (the lamp pass). The "M1" in the header is the choice you made in the filter row, for Build Profile. M1 and M2 themselves are in other files of the run (see below), so I could not check them yet.
- The readings look sound: no negative or clipped values, smooth spectra, paper white at L* 94.5 to 94.9, black at L* 5 to 6.
- Your paper has an optical brightener. Its white reflects about 93 % at 430 to 440 nm, more than the 86 % around 570 nm, and only 38 to 39 % at 400 nm. That rise in the blue is the typical fingerprint of a brightener. On this paper M1 and M2 should therefore come out clearly different from M0.
- Scan against rescan, two separate measurements each with its own calibration: leaving strip A aside, the 684 patches differ by 0.17 Delta E 2000 (median), 0.21 on average, and only 5 by more than 1.0 (the largest is 1.3, on a near black). That is about what we would expect from an i1Pro 2 read by hand twice.
One thing to look at: strip A in the second file
The first three patches of strip A in your second file do not match the chart. A2 should be orange and reads as a grey-green. A1 to A3 differ from your first file by 30 to 43 Delta E 2000, and A4 by 4. From A5 on, the strip agrees with your first file. It looks like a misread at the start of that strip, in the lamp pass. Your first file does not have it, so for now that is the one to build a profile from.
What would let us check M1 and M2
From each of the two runs (in your ChromIQ folder: your project, then "runs", then run1, run2 and so on), please send the folders "conditions" and "dual_scan". "conditions" holds M1 and M2 (files ending in -M1.ti3 and -M2.ti3), "dual_scan" holds ChromIQ's record of both passes of every strip; it contains no serial number. The text of the Measure log helps too. With these we can compare M0, M1 and M2 patch by patch, see whether every strip kept its M1 and M2, and compare your two runs in all three conditions.
The lights
The green and red blinks come from the other Beta option, "show the instrument's state with its lights", with result signals on. ChromIQ only sends light sequences it has seen X-Rite's i1Profiler send, in USB recordings, and only where i1Profiler sent them. The recordings we have of i1Profiler doing a dual scan with an i1Pro 2 show what it does: after the first pass a blue pulse on the left side, and green or red only after the return pass. In a row whose first pass had failed, i1Profiler's red also came only after the return pass. ChromIQ has the same reason to wait: the return pass is also what ends the instrument's position reading in the way the recordings show, so it is needed even after a bad first pass.
The blue pulse is the part ChromIQ does not send yet. It is in those recordings, so it would be allowed. The difficulty is timing: i1Profiler sent it about half a second to one second before the return pass, while ChromIQ keeps the instrument's lights quiet for 2.1 seconds before every reading, since an i1Pro 2 lost readings when a light command came just before one. The recorded pulse also always lights the left side, while ChromIQ lets you read the return pass in either direction. We will look into whether the blue pulse can be fitted in safely, or whether a sound or the Measure tab can show the return pass more clearly. I cannot promise which yet.
Your watercolour paper
That is a very good check. M0, M1 and M2 differ only in how much UV light reaches the paper, and without a brightener the UV has almost nothing to act on. So on paper with little or no brightener, the three should come out nearly the same, and its white should show no rise around 430 to 440 nm. If they come out clearly different there, something is wrong on our side. No paper without brightener has been measured with the dual scan yet. When you get to it, the "conditions" and "dual_scan" folders of that run are what tell us most.
Thank you again. This was a big step for the dual scan.