Measuring the ColorChecker Classic with the i1Pro 2/3 Scanning Ruler - Part 1

. Developing a Measurement Method for the ColorChecker Classic Using the i1Pro 2/3 Scanning Ruler
. Establish a Reference: Obtain reference measurements using Spot Measurement.
. Perform Scanning Measurements: Measure the ColorChecker Classic 24 using the i1Pro 2/3 scanning ruler.
. Determine the Compensation Factor (CF): Compare the scanning measurements against the Spot Measurement reference to calculate the required Compensation Factor

Why the Raised Paper Frame Causes Problems
The physical ColorChecker Classic 24 has a raised paper frame surrounding its recessed color patches.

When using the i1Pro 2/3 scanning ruler, the ruler rests on the raised frame rather than the actual measurement surface.
This increases the distance between the instrument's measuring head and the color patches, potentially affecting measurement accuracy.

Precautions When Taking Spot Measurements
Position the target aperture precisely over the centre of each color patch.

Press the i1Pro 2/3 straight down firmly against the positioning target when taking each reading. Ensure that the measuring surface is as close as possible to the instrument's measuring head.


Measurement Workflow
- Perform three sets of measurements using Spot Measurement.
- Perform three sets of measurements using the scanning ruler.
- Calculate the average results for both Spot Measurement and scanning ruler measurements.
- Determine the Compensation Factor (CF) by comparing the two sets of averaged results.
- Apply the calculated CF to the scanning ruler measurement results.

1795841438044.png



Method A — Calculation Method
Method B — Wavelength-dependent Compensation
 
Measuring the ColorChecker Classic with the i1Pro 2/3 Scanning Ruler -Part 2

Method A: Scalar CF

This method applies one constant compensation factor to every patch and every wavelength
Model complexity -Simpler, same multiplier throughout, preserved Spectral curve shape

. One factor for the whole spectrum.
. corrected = CF x measured

Method B: Wavelength CF + wavelength shift
Model complexity -36 factors, varies by wavelength, modified Spectral curve shape

. corrected(l) = CF(l) x measured(l) + S x slope(l)
. 36 factors (one per wavelength) plus ONE wavelength-shift term S.
. slope(l) is how steeply the measured spectrum rises or falls at l.
. All 37 values are fitted together by least squares.

Averaging Measurements

Create Reference Set: ColorCheck24-Spot-Avg-Reference.mxf
ColorCheck24-Spot-set1.mxf
ColorCheck24-Spot-set2.mxf
ColorCheck24-Spot-set3.mxf

Measurements using the scanning ruler, create ColorCheck24-Ruler-Avg.mxf
ColorCheck24-Ruler-set1.mxf
ColorCheck24-Ruler-set2.mxf
ColorCheck24-Ruler-set3.mxf


Scalar CF vs Wavelength CF

Before CF:

ColorCheck24-Spot-Avg-Reference.mxf vs ColorCheck24-Ruler-Avg.mxf
1795911424783.png


After Scaler CF:
ColorCheck24-Spot-Avg-Reference.mxf vs ColorCheck24-Ruler-Avg_CF-Scalar-1.012474.mxf
1795911437883.png


After Wavelength CF:
ColorCheck24-Spot-Avg-Reference.mxf vs ColorCheck24-Ruler-Avg_CF-WavelengthShift.mxf
1795911451976.png


Interpretation
The comparison revealed a maximum ΔE00 difference of just 0.08 between the two corrected methods.
This demonstrates an exceptionally close agreement in their colorimetric results, despite the use of different spectral correction approaches.
 