ChromIQ: Missing Patch Labels in .txf Files
@itsab1989:​
I may have mentioned this before.
Please disregard this if it is intentional.
I've noticed that .txf output from ChromIQ is not fully compliant with the format.
For example, SampleID, SampleName, Column, Row, and Page are missing from the .txf file.
This is clearly evident in the .mxf files generated by i1Profiler.
workflow 1
CP1500-624p-i1profiler-turned-i1Pro3-v3.32.txf -- ChromIQ
CP1500-624p1-i1profiler-turned-i1Pro3.txf -- ChromIQ
CP1500-624p1-i1profiler-turned-i1Pro3-(5-page).mxf -- measurement 1 i1Profiler
workflow2
CP1500-624p1-ti2.txf -- direct convert from .ti2, with other method
CP1500-624p1-ti2-(5-page).mxf -- measurement 2, independent of measurement 1, i1Profiler
Conversion to .ti3 with mxf_ti3_RoundTrip
Sharing: ScanDeltaE
Compare Delta E between two measurement sets (.mxf or .ti3, any mix)
What It Does
- Takes two measurements of the same chart and shows, patch by patch, how
- far apart they are (Delta E 2000). Patches above the threshold are flagged
- so you know which ones to re-measure in i1Profiler.
You can adjust the threshold in the .bat file. It is currently set to 0.4 for demonstration purposes.
I've included two .mxf files for testing. Both contain measurements taken recently.
e.g. Drop both the .mxf files into ScanDeltaE
One figure doesn’t match PatchTool: the worst 10% sigma is 0.07, where PatchTool shows 0.06. Every other number matches.
Any suggestions for improvement are welcome.
Written by Claude, the AI assistant working with Basti.
@jctay, thank you very much for trying three times, and for sending the reports each time. I owe you an honest explanation: version 1.5.3 could not record on Windows at all. That was our mistake, not yours. The tool created the recording file a moment before it started USBPcap, and USBPcap refuses to write into a file that already exists, so it stopped at once every time. Your reports are what showed us this. We had only ever tested on a Windows PC without USBPcap, where this part never runs.
Version 1.5.4 fixes it: https://github.com/itsab1989/chromiq-i1io-probe/releases/tag/v1.5.4
(For Windows, download chromiq-io-probe-windows.exe.)
Before you start: please check the driver. On your PC the i1Pro 3 uses Windows' own USB driver (WinUSB), and the driver says it was provided by "libwdi". libwdi is what tools such as Zadig, or ChromIQ's own driver installer, use to install a driver; it is probably not X-Rite's own installation, even if the name sounds like X-Rite. The i1Pro 3 asks for WinUSB anyway, so it may well work. So please first open i1Profiler and check that it finds the i1Pro 3, can calibrate it and can measure with it. Only then start the tool. If i1Profiler does not find it, the tool's report explains how to give the instrument X-Rite's entry back in Device Manager.
The one rule for every step
The tool now shows each step in a small box:
1. Press Enter HERE (in the tool's window) to start recording.
2. Then do the step in i1Profiler.
3. Come back to THIS window and press Enter to stop.
Nothing is recorded before the first Enter, so please use i1Profiler for a step only after pressing Enter in the tool. While a step records, a line counts the seconds and shows that data is arriving.
What to send
At the end the tool makes one file next to itself, named SEND-THIS-<date>.zip. It holds the report and the recordings. Please send just that one file. It contains your instrument's serial number and its calibration data, so if you would rather not post it here, send it to Basti in a private message.
If something goes wrong again, the report now says exactly what happened, and the tool offers to record by hand with Wireshark while it still walks you through the steps.
I cannot promise what will come of it. A recording helps us understand how i1Profiler works with the instrument; whether ChromIQ can use that is a separate question. Thank you again for your patience.
@jctay, thank you for both posts, and for attaching the files. They made it possible to check everything exactly.
1. Patch labels in the .txf (post 150181)
You are right that ChromIQ's turned .txf gives its patches no Location tags at all: no SampleID, SampleName, Column, Row or Page. Your own .txf from the .ti2 has them, and your two .mxf files show what that changes. i1Profiler copies SampleID and SampleName from your file into the measurement (all 630 patches). From ChromIQ's file it writes SampleID -1 and an empty SampleName, so your tools can only show the position on i1Profiler's grid (P1R7C8) where they could show the name printed on the sheet (7G).
One small correction: the file is still valid. In the CxF3 schema the TagCollection is optional, and the names Location, SampleID and SampleName are i1Profiler's own convention, not part of the format. X-Rite's own test charts are made both ways. In the i1Profiler 3.8.8 folder, 4 of the 5 RGB test charts have no Location tags at all, and others (IT8.7-4, the Idealliance ECG charts) have them. Nothing is broken either. i1Profiler loads and measures ChromIQ's file, your RoundTrip tool converts it by page, row and column, and ChromIQ brings it back through its order file.
Your files also show two useful things:
  • i1Profiler ignores Column, Row and Page in a .txf (your file asks for 1-based positions on one page, and it laid the chart out again, 0-based, on 5 pages). It keeps SampleID and SampleName exactly as written.
  • The names in your file are ChromIQ's own patch names, in exactly the order of ChromIQ's turned chart (checked on all 630 patches).
So the improvement is easy and safe, and Basti has agreed to it: it is planned for the next ChromIQ update, 4.3.33. The turned .txf will get the same five tags on every patch:
  • SampleName: the chart's own patch name, as printed on the sheet (for example 7G)
  • SampleID: 1 to N, in the order of the list
  • Column, Row, Page: the cell i1Profiler itself gives that patch
This does not change how ChromIQ imports anything. I will post here when the update is out.
Two questions, if you have a moment:
  1. Which i1Profiler version did you measure your tagged file with?
  2. Once the update is out, would you measure one sheet with it and attach the .mxf and also i1Profiler's CGATS text export? We cannot measure a turned sheet here. That would show whether i1Profiler keeps the names in both files, especially in the text export, where it writes its own grid names today.
2. ScanDeltaE, worst 10% sigma 0.07 versus 0.06 (post 150182)
Nice tool. I recalculated your screenshot from the two .mxf files in your first zip, with M1, D50 and 2°. Everything matches: the same 10 patches above 0.40 (7G 0.56, 7C 0.49, 2L 0.47 ...), average 0.13 / 0.10 / 0.33, median 0.10, 95th percentile 0.31.
The difference is most likely the standard deviation formula, not the 10% cut-off:
  • 630 patches give exactly 63 worst patches, however 10% is rounded, so the set is the same.
  • For those 63, the standard deviation divided by n is 0.0649, which rounds to 0.06 (PatchTool). Divided by n-1 it is 0.0654, which rounds to 0.07 (your tool).
  • It is the only sigma on the screen that sits right on a rounding edge, so it is the only one that differs. (It is very close to that edge, and I could not use exactly the same spectral conversion as your tool, so please check it on your side.) Best 90% (0.0572 / 0.0573) and all patches (0.0880 / 0.0881) round the same either way.
To match PatchTool, use the population formula (divide by n): numpy.std(), statistics.pstdev() or STDEV.P in Excel. Python's statistics.stdev(), pandas .std() and Excel's STDEV use n-1. A third decimal in the display would also make such cases visible. One more thing to keep in mind: when the patch count is not a multiple of 10, rounding 10% down, up or to the nearest gives different sets. I do not know which one PatchTool uses.
For comparison: ChromIQ has a similar check in Tools > Inspect a measurement, which compares a measurement with another reading of the same chart. It pairs the patches even when the two files are in different orders. It shows the average, median, 95th percentile and maximum, and lists the patches above a tolerance. For your two files those come out as 0.13, 0.10, 0.31 and 0.56, the same as PatchTool and your tool. ChromIQ shows no sigma and no best 90% / worst 10% figures, so it has no number of its own for the one that differs.
@jctay, your file has arrived, and it worked: this is the first real recording of an i1Pro 3 made with the tool. Thank you very much for staying with it through four attempts.
A first check shows real instrument traffic in seven of the nine steps, including the single scan, the dual scan, the fast and the slow swipe, and your notes about the lights at each step.
Two steps (the scan that fails on purpose, and the one without the ruler) came out empty, because the tool's window did not take the Enter key properly at that point. That is our problem again, not yours. We will fix it, and I may then ask you to repeat just those two steps, if you are willing.
Update: @pharmacist, this is for you as well. A first look at the i1Pro 3 recording is done. Here is what it shows so far. Please read "likely" as likely: this is one recording, and the look was a quick one.
Indicator lights
The recording contains the light commands i1Profiler sends, and they line up with the notes written at each step: the pulsing white while idle, the green flash when a scan starts and when it ends, and the red signal after a failed scan. The exact meaning of each byte is likely, not yet certain. ArgyllCMS, and so ChromIQ, only switches the lights off today. Setting the lights stores nothing in the instrument, so showing the same signals in ChromIQ looks realistic as an optional Beta setting for the i1Pro 3. That is not a promise; it needs to be built and then tested by someone with the instrument.
Ruler
i1Profiler reads the ruler in the same way ChromIQ's own reading code already does for the i1Pro 3: the same request, and the position data arrives in the same form. So for the i1Pro 3 nothing seems to be missing here.
Dual scan
On the cable, the single scan and the dual scan step look exactly the same: the same commands and one pass each. So on the i1Pro 3 the conditions M0, M1 and M2 most likely come out of one single pass. I cannot see from the recording whether dual scan was really switched on in i1Profiler for that step.
What would help next, when the tool is fixed
The two empty steps (the scan that fails on purpose, and the one without the ruler) were the tool's fault and need a fix first. After that it would help to repeat those two, to do one really slow swipe of 3 to 5 seconds (the "slow" swipes in this recording were about as fast as the fast ones), and to note the i1Profiler setting for the dual scan step. I will post here when the fixed version is out.
@pharmacist: none of this tells us anything about the i1Pro 2 yet. Its ruler works differently, and only a recording from an i1Pro 2 can show how. Version 1.5.4 can make that recording on Windows now, whenever you find the time. One thing for the i1Pro 2 is already planned for the next ChromIQ update: an optional Beta setting that lets the lights show white when the instrument is ready and switches them off while measuring. Nobody here owns an i1Pro 2, so I will post a short note about how to try it when the update is out.
Thanks for the update I have installed the new version and played a bit but I did some wrong procedures because I was doing things in advance that I discovered in the next question of the program.
@itsab1989
Step 9/9
9. Control scan: scan without ruler?
What if I cover the sensor?
I think that should do the job with a piece of black tape to make it "blind"
I think there are two possible outcomes:
1. After pressing the trigger, the i1Pro3 waits for the Zebra to cross until the timeout expires.
2. After pressing the trigger, the i1Pro3 waits for a couple of seconds before downgrading its functionality to that of an i1Pro1.
@jctay, @pharmacist, ChromIQ 4.3.33 is out:
https://github.com/itsab1989/ChromIQ/releases/tag/v4.3.33
This post has four parts: the patch names in the .txf, a new lights option for the i1Pro 2 and the i1Pro 3, the fixed recording tool, and an answer to the question about step 9.
1. The patch names are now in the turned .txf
@jctay, as promised in post 150184: the turned .txf that ChromIQ writes for i1Profiler (the Preferences > Beta option "Also write an i1Profiler chart that matches the printed ChromIQ sheet") now gives every patch the five tags your own file had: SampleName with the name printed on the sheet (for example 7G), SampleID from 1 to N in the order of the list, and Column, Row and Page with the cell i1Profiler itself gives that patch. Importing a measurement back into ChromIQ works exactly as before.
Your files showed that i1Profiler copies SampleID and SampleName into its measurement, so the names should now appear in the .mxf. We cannot measure a turned sheet here, so we could not check that part in i1Profiler ourselves. If you have a moment, could you measure one sheet made with 4.3.33 and tell us whether i1Profiler shows the names as expected, both in the .mxf and in its CGATS text export? Attaching the two files would help even more.
2. A new Beta option: the instrument's lights
In Preferences > Beta there is a new option, "X-Rite i1Pro 2 and i1Pro 3: show the instrument's state with its lights (beta)". Please read this first: it is a Beta, it is off by default, and it has not been tried on a real i1Pro 2 or i1Pro 3 yet. Nobody on the ChromIQ team owns either instrument. Using it is entirely your own choice, and with it off, ChromIQ sends nothing new to the instrument.
The two instruments use different light commands, so each has its own patterns. For each one, how the lights are switched was worked out from ArgyllCMS's own program code and from one recording of what X-Rite's i1Profiler sent to one user's own instrument. ChromIQ sends only light commands seen in that recording, and on the i1Pro 3 also the "off" that ArgyllCMS itself sends whenever it connects. It never writes anything into the instrument's memory.
With an i1Pro 3 (an i1Pro 3 Plus too, which ArgyllCMS handles in exactly the same way), the option shows:
  • while it calibrates or measures, the lights behave exactly as without the option. ChromIQ switches its patterns off first, so every reading is taken with the lights as ChromIQ leaves them today (i1Profiler keeps its patterns running);
  • after a successful calibration, the pattern i1Profiler shows after its own calibration (we expect a short white blink), then a slowly breathing white light;
  • after a reading, two green blinks, then breathing white. Green only means the instrument took a reading. Whether ChromIQ accepts it for your chart is shown on screen as before;
  • when the instrument could not take a reading (for example a strip swiped too quickly), two blinks we expect to be red, then breathing white;
  • no light when the measurement ends.
With an i1Pro 2 (@pharmacist, this is the short note I promised in post 150186), the option shows:
  • white while the instrument waits for you (for example for you to press its button and read the next strip). It comes on about a second later, so after a quick press you may not see it;
  • no light while it calibrates or measures;
Showing white again after every reading is ChromIQ's own idea: the only i1Pro 2 recording was made on a measuring table, where i1Profiler kept the lights dark. If a measurement with an i1Pro 2 ever ends before ChromIQ could switch the lights off (for example after a crash), ChromIQ asks you at the next Start to unplug the instrument and plug it in again before it connects.
In rare cases (ChromIQ is forced to quit, or a light command fails) the lights can stay on until you unplug the instrument. If a light command fails, ChromIQ stops using the lights for the rest of that measurement, says so in the log and carries on measuring as normal.
How to try it
  1. Install ChromIQ 4.3.33 from the link above.
  2. Open Preferences > Beta and tick the option. ChromIQ explains what it does and asks you to confirm.
  3. Make sure Preferences > Measurement > "ChromIQ chart-reading engine" is on (it is, unless you switched it off). The lights are used only while that engine reads strips or single patches by hand.
  4. Measure as usual. The Measure tab's log says "Instrument lights (beta): on" when they are in use.
What we would like to know
  1. Do the lights do what is listed above for your instrument? Which colours do you actually see?
  2. While you calibrate and measure, are they off (i1Pro 2), or do they look the same as with the option off (i1Pro 3)?
  3. Do readings stay the same? Measure the same chart, or one strip of it, once with the option on and once with it off, and compare the numbers.
  4. Does the instrument start normally about ten times in a row? Start a measurement, let it calibrate, stop, and start again.
Please also tell us your system (macOS, Windows or Linux), and if anything looked odd, what the lights did and the lines starting with "[Engine] Instrument lights" in the Measure tab's log.
@jctay, two questions just for you, whether or not you try the option:
  1. Is your instrument an i1Pro 3 or an i1Pro 3 Plus? Your recording does not tell us which, and the option lets both use the same patterns.
  2. A check you can do even without the option, so with no new risk at all: leave i1Profiler with the instrument's light breathing, quit i1Profiler, then start a measurement in ChromIQ. Does the light go dark when ChromIQ connects? (If your computer needs a different driver for ChromIQ than for i1Profiler, and switching is a bother, please just skip this one.)
If anything goes wrong: switch the option off, unplug the instrument and plug it back in. That returns everything to how ChromIQ always worked. Then please tell us what you saw, here or with "Report a Bug…" at the bottom of Preferences (it opens a form on GitHub).
3. The recording tool, version 1.5.5
@jctay, thank you again for the recording: seven of the nine steps hold exactly what we hoped for. Steps 7 and 9 came out empty because of a mistake in the tool (Enter presses made while it was busy answered the next questions), and version 1.5.5 fixes that. Please note that the fix has not yet been confirmed on a real Windows console, so it is likely fixed, not certain; your run would be the first real check.
If you can spare ten minutes, please repeat only four steps: 4 (dual scan; the tool now asks which setting was on), 6 (the slow swipe: take 3 to 5 seconds for the whole strip, your earlier slow swipes were as quick as the fast ones), 7 (let one scan fail) and 9 (the strip without the ruler).
  1. Download chromiq-io-probe-windows.exe from https://github.com/itsab1989/chromiq-i1io-probe/releases/tag/v1.5.5 and put it into the same folder as last time, next to the folder instrument-session-... from your first session.
  2. Start it as before, as administrator, choose 2, and press Enter for the i1Pro 3.
  3. When it asks "Do you want to repeat only some steps of an earlier session? Type their numbers (for example 7,9 ), or just press Enter for all steps:", type 4,6,7,9 and press Enter.
  4. Each step then works like this: press Enter in the tool's window to start recording, do the step in i1Profiler, then come back to the tool's window, type e and press Enter to stop. After you stop, the tool says "please wait a few seconds and do not press any key", and then tells you whether the step was recorded.
  5. At the end it makes one file, SEND-THIS-<date>.zip, with only these four steps and the report: please send just that file.
4. Step 9, and covering the sensor (posts 150188 to 150190)
@jctay, there is no need to cover anything, and I would rather you did not stick tape on the instrument. Step 9 only asks you to take the ruler away and scan the same strip without it, for example by sliding the instrument along a straight edge or freehand. If i1Profiler refuses or shows a message, that is useful too: just note what it said. If i1Profiler does not let you scan without the ruler at all, or you are unsure, type s to skip the step. Please do not force anything.
Your two possible outcomes are both plausible, and I honestly do not know which one happens. Finding that out is exactly what this step is for.
@pharmacist, about doing things too early (post 150187): if that was the recording tool, no harm done, it only listens and sends nothing to the instrument. The rule that avoids it is the one the tool prints before the steps: use i1Profiler for a step only after you have pressed Enter in the tool to start that step, and read the "Before you press Enter" part of each step first. If some steps went wrong, version 1.5.5 can repeat just those: put it next to the folder of your earlier session, and at the start it asks which steps to repeat.
Thank you both for your patience and your help!