@jctay, thank you, the three .mxf files read without trouble. I calculated Lab from the spectra as before (ArgyllCMS spec2cie, D50, 2 degree observer; differences in dE00). The files hold M0, M1 and M2; the three conditions tell the same story, so I give M1 and note where another differs.
Chart hd2, patch length 8.7 mm (about 8.4 as printed), i1Profiler only
Repeatability, set 1 against set 2, all 196 patches:
  • mean 0.09, 95 % of the patches below 0.27, largest 0.48 (patch 8D).
That is as good as on the 7.4 mm chart (0.09 there as well).
Row 11, the strip readings against your spot readings (14 patches):
  • set 1: mean 0.65, largest 1.28. Set 2: mean 0.65, largest 1.27.
  • Direction: the strip reading is lighter than the spot reading on 13 of the 14 patches, by 0.3 to 0.4 L* on average.
  • Largest differences: 11E (1.3), 11C, 11G, 11L and 11M (each about 1.0 to 1.1).
Next to the 7.4 mm chart (row 12 there):
patch lengthstrip against stripstrip against spot, meanlargeststrip lighter by
7.4 mm0.090.741.480.6 L*
8.7 mm0.090.651.280.3 to 0.4 L*
What I read from this, and what I do not:
  1. Going from 7.4 to 8.7 mm made the strip reading a little closer to the spot reading, but only a little: 0.74 to 0.65 on average. It did not bring the two together.
  2. The difference has the same direction on both charts (strip lighter than spot), and it shrank by about a third. That part fits the idea that neighbouring patches leak into a strip reading, less so as the patches get longer.
  3. But a difference of 0.65 that remains at 8.7 mm makes me doubt that all of it is the patch length. Part of it may be a plain difference between a spot reading and a strip reading in i1Profiler that would also be there on large patches: the instrument rests still on one spot in one case and averages along a moving path in the other, and a print is never perfectly even inside a patch. From these two charts I cannot tell the two causes apart.
So before drawing a line for the minimum length, I think the reference itself needs one check: the same comparison, strip against spot for one row, on a chart with clearly large patches (12 mm or more along the strip). If strip and spot agree there to about 0.2, then the 0.65 at 8.7 mm is the patch length and the search continues upward (9, 9.5, 10 mm). If they still differ by about 0.5 there, then spot readings are not the right yardstick for strip readings, and the better reference would be a large-patch strip reading of the same colours.
Two caveats as before: one row of 14 patches each time, and the two charts have different patch sets, so row 11 here and row 12 there are different colours.